Vezi aici Testing diacs and sidacs https://www.repairfaq.org/sam/semitest.htm#sttds sau aici http://www.edaboard.com/thread227947.html

Sau poti sa incerci si asta:

 



The 100V SIDAC (for the series reactor ballasts for the 55V HPS lamps) you may quite easily check as well:
1) An Ohm-meter should read an open circuit
2) Connect a regular incandescent bulb (20..40W) in series with the SIDAC to the mains, the lamp should light at a reduced brightness, but without any 60Hz flicker. If you own a variac, you may try to connect the lamp to the output and observe the response. The lamp should remain completely OFF from zero till about 70..80V, then should rather abruptly light up to reduced brightness. A small region (5..10V) where the lamp is flickering at 60Hz (only one polarity conducting) is OK.

The 200V SIDAC's (for the lamps with arc voltage of 70V and above, so S56 and higher wattages) could be tested principally the same way, you only need all voltages double. A 120/240V tapped ballast primary used as a step up transformer and either 240V or two 120V bulbs in series would do the job rather safely (for a lab test purpose).

Trebuie sa ai grija ca la testari sa nu depasesti curentul maxim suportat de 1A.Bine e,cum se zice mai sus,sa ai un autotransformator.

Link to comment

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now



×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.Terms of Use si Guidelines